HP 6890/5972A GC-MS: "cannot bring up GC" error [Copy URL]

Rank: 8Rank: 8

Post time 2013-2-3 12:40:43 |Show all posts

Hi everybody,

I am using an HP 6890/5972A GC/MS w/ Chemstation GC1701BA version B.01.00. The software was recently reinstalled and now will not connect to the instrument. The interface is an Agilent 82350B PCI GPIB. Thanks in advance for any help you can provide!

-Pat

Use magic Report

Rank: 8Rank: 8

Post time 2013-2-3 12:40:45 |Show all posts

Did you configure the GPIB board as "hp82341" even though it is an 82350B ? You need to. And change the bus address to 30. You need to.

Use magic Report

Rank: 8Rank: 8

Post time 2013-2-3 12:40:47 |Show all posts

Thanks very much, that fixed it. The problem was an incorrect SICL name and bus address. You guys are a great help!